Intermediate precision validation: how many instruments?
Posted: Wed Mar 21, 2007 4:38 pm
In my company we have some HPLC instruments, all the same brand and model.
Usually, when I perform validations studies I include two different HPLC systems as possible source of variation (together with analyst, day, column, etc).
I have always assumed that when you have proved that results do not change between the two instruments, you can extend the assumption to all other HPLC systems, at least if you don't change the instrument model.
Recently, an external auditor seemed to disagree with this point of view, requiring a confirm that release analyses were performed on the same instrument where the validation was performed.
Now I am curious: is this point a common opinion?
Because it seems to me quite unpractical either to include all the HPLC systems of the laboratory in a validation study, or "to lock" the possibility to perform an analysis only on those two instruments where the validation was performed.
Thank you for opinion
Usually, when I perform validations studies I include two different HPLC systems as possible source of variation (together with analyst, day, column, etc).
I have always assumed that when you have proved that results do not change between the two instruments, you can extend the assumption to all other HPLC systems, at least if you don't change the instrument model.
Recently, an external auditor seemed to disagree with this point of view, requiring a confirm that release analyses were performed on the same instrument where the validation was performed.
Now I am curious: is this point a common opinion?
Because it seems to me quite unpractical either to include all the HPLC systems of the laboratory in a validation study, or "to lock" the possibility to perform an analysis only on those two instruments where the validation was performed.
Thank you for opinion