Is that normal? It strikes me as odd, but perhaps I'm missing something.
Belle Chasse, LA
Advertisement
Discussions about GC and other "gas phase" separation techniques.
It's just that the FID has ca. 450 cc/min of gas flowing through it, between the fuel, air and make-up. I know that you can get diffusion against gas flow, but the effect seemed rather dramatic.I would say that this problem is 'normal', but it could also indicate that there is a problem woth the FID. If you think about it, if you can smell the acetone, then it is in the air around the GC, and therefore it will diffuse down into the FID and it will be burnt. You could also have a sealing problem with the upper part of the FID, and it could be possible that the acetone is being drawn into the FID through a badly sealing 'O'ring.
I have seen similar problems with chlorinated solvents close to an ECD, the stuff just diffuses up the vent line when customers do not connect the vent of the ECD to a proper ECD vent line