Misinjection problem
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
6 posts
Page 1 of 1
The HPLC system I am using has an autosampler and has a high frequency of misinjection. Is there anything i can do to fix it?
Peter
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
Let me be the one to have the voice of reason:
What brand of HPLC/autosampler?
What model number?
How old is it?
What does "high frequency of misinjection" mean? doesn't pick up vial, doesn't suck up any sample, to quote Yul Bryner "etc., etc., etc."
What brand of HPLC/autosampler?
What model number?
How old is it?
What does "high frequency of misinjection" mean? doesn't pick up vial, doesn't suck up any sample, to quote Yul Bryner "etc., etc., etc."
-
- Posts: 3
- Joined: Wed Jan 25, 2006 3:57 pm
A couple of other things you can check are:
By mis-injection do you mean you don't get any peaks or variable peak sizes?
Does anything else change in the chromatography on the mis-injections i.e. retention times or baseline profile?
Is there anything in the HPLC electronic log that is different between a 'good' and a bad' injection?
What is the sample solvent?
By mis-injection do you mean you don't get any peaks or variable peak sizes?
Does anything else change in the chromatography on the mis-injections i.e. retention times or baseline profile?
Is there anything in the HPLC electronic log that is different between a 'good' and a bad' injection?
What is the sample solvent?
Mike Scott
-
- Posts: 179
- Joined: Tue Nov 09, 2004 5:04 pm
In addition to what has already been mentioned, you can include a check of the autosampler vials.
Are you using the vials recommended by the manufacturer? Or at least are you using an appropriate/equivalent substitute vial? Some autosamplers are not forgiving if the wrong vial is used and missed/skipped/bad injections can result.
Regards,
Dan
Are you using the vials recommended by the manufacturer? Or at least are you using an appropriate/equivalent substitute vial? Some autosamplers are not forgiving if the wrong vial is used and missed/skipped/bad injections can result.
Regards,
Dan
-
- Posts: 3
- Joined: Wed Mar 14, 2007 2:23 am
thans for your reply.
It's Agilent 1200 series model G1329A. It's a new one. It draw less volume so getting a low peak area. In a batch of 40 to 50 samples it would happen once or twice. I don't think it's normal compared to other HPLC we have. I have tried to change the draw and eject speed, But it didn't help a lot. Any advice?
It's Agilent 1200 series model G1329A. It's a new one. It draw less volume so getting a low peak area. In a batch of 40 to 50 samples it would happen once or twice. I don't think it's normal compared to other HPLC we have. I have tried to change the draw and eject speed, But it didn't help a lot. Any advice?
Peter
-
- Posts: 5
- Joined: Thu Jan 25, 2007 6:46 pm
Since you didn't mention adjusting it, I'll mention the obvious - the draw position. Just below the settings for the draw and eject speeds is the draw position. I think the default is 0.0mm which should pick up sample from about the 0.2mL position in the autosampler vial. If you have very limited volume of sample, you can lower the draw position as low as -2.5 mm, but verify that the injection needle doesn't touch the bottom of the autosampler vial.
And if you have a new 1200, it should be still under warranty (1 year?), call 1-800-227-9770, option #3.
And if you have a new 1200, it should be still under warranty (1 year?), call 1-800-227-9770, option #3.
Kathy
6 posts
Page 1 of 1
Who is online
In total there are 3 users online :: 2 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Baidu [Spider], Google [Bot] and 1 guest
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Baidu [Spider], Google [Bot] and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science