Amino Acid derivatization
Posted: Sat Mar 17, 2007 12:22 pm
Hello to everyone.
I'm very new to LC analyzes, and now I'm trying to create a method for the analyzation of protein hydrolysates.
Since I don't have a fluoresence detector, I'm stuck with the PITC-method. I created my very first standards, from a standard solution of 1mM (of each AA) in 6M HCl. I chose this, because I'm going to do the hydrolysis with the HCl method, so I thought the standard should be in the same matrix.
But alas... when I did the first run, I had very strange results (read: only like five clear peaks and the standard should contain 20 aa), and now I don't know where to start the problem solving. I diluted my standard in 1:100. Here are my LC setup and the method I used for the derivatization of the AA standard:
Restek Pinnacle DB c18 - column, in +40c
flowrate 1 ml/min
Detector at 254nm
Buffer gradient from A to B, in 25min
Buffer A:
38g Na-Acetate
2l mQ-H2O
1,1ml TEA (Triethylamine)
Buffer B:
60% Acetonitrile
Derivatization:
Take 25ul of your sample in HCl (standard in this case) and dry it using Nitrogen gas (I don't have a vacuum exicator...). Resuspense the sample into EMT (Ethanol, mQ, TEA 2:2:1 vol/vol/vol) and dry it again. Repeat. Resuspense the sample into freshly made PEMT (PITC, EtOH, mQ, TEA, 1:7:1:1 vol/vol/vol/vol). Let it react in rt for apprx. 20min and dry it again. Resuspense the sample into 400ul of Buffer A and transfer the sample solution into a 1,5ml vial and seal it. Begin HPLC analysis
I'd be very happy if you could point me to my flaw, or to the direction where to look for problems. And what would you recommend for the consentration of the standards?
Thank you in advance
Heikki Laine
I'm very new to LC analyzes, and now I'm trying to create a method for the analyzation of protein hydrolysates.
Since I don't have a fluoresence detector, I'm stuck with the PITC-method. I created my very first standards, from a standard solution of 1mM (of each AA) in 6M HCl. I chose this, because I'm going to do the hydrolysis with the HCl method, so I thought the standard should be in the same matrix.
But alas... when I did the first run, I had very strange results (read: only like five clear peaks and the standard should contain 20 aa), and now I don't know where to start the problem solving. I diluted my standard in 1:100. Here are my LC setup and the method I used for the derivatization of the AA standard:
Restek Pinnacle DB c18 - column, in +40c
flowrate 1 ml/min
Detector at 254nm
Buffer gradient from A to B, in 25min
Buffer A:
38g Na-Acetate
2l mQ-H2O
1,1ml TEA (Triethylamine)
Buffer B:
60% Acetonitrile
Derivatization:
Take 25ul of your sample in HCl (standard in this case) and dry it using Nitrogen gas (I don't have a vacuum exicator...). Resuspense the sample into EMT (Ethanol, mQ, TEA 2:2:1 vol/vol/vol) and dry it again. Repeat. Resuspense the sample into freshly made PEMT (PITC, EtOH, mQ, TEA, 1:7:1:1 vol/vol/vol/vol). Let it react in rt for apprx. 20min and dry it again. Resuspense the sample into 400ul of Buffer A and transfer the sample solution into a 1,5ml vial and seal it. Begin HPLC analysis
I'd be very happy if you could point me to my flaw, or to the direction where to look for problems. And what would you recommend for the consentration of the standards?
Thank you in advance
Heikki Laine