Glufib (standard) coming off column far too late! Help!

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
I am using LC Packings Ultimate-Switchos-Famos coupled to Applied Biosystems QStar Mass Spec. I believe it is a typical set-up for most people doing shotgun proteomics.

Unfortuntely, our standard which is Glufib is coming off far too late in a 30 min gradient. The pressures are fine and a colleague has suggested we have dead volume in the system. Could you suggest a way we can sort this problem out and whether it is dead volume issue?

Thanks for your time

Euhalo-Jags
Often also shows in poor peak shape. If you have tailing on the peak then you may well have deadtime. Check all your fittings!

Hi, thanks for the reply. We had to take the valve and all the fittings apart, give them a clean with methanol and then it worked.

many thanks for your help.
3 posts Page 1 of 1

Who is online

In total there are 28 users online :: 2 registered, 0 hidden and 26 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Amazon [Bot] and 26 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry