6890 micro ecd's
Discussions about GC and other "gas phase" separation techniques.
2 posts
Page 1 of 1
We have several HP5890's where we have had good success reaching our detection levels for chlorinated pesticides, toxaphene and pcb's. We have acquired some Agilent 6890 GC's equipped with micro-ecd's and can't seem to replicate the DL's achieved on the HP5890's. We are using Turbochrom for data collection/reduction. Any suggestions?
-
- Posts: 176
- Joined: Mon Mar 14, 2005 7:01 pm
Hi!
6890 micro ECD is more sensitive than 5890....very more!
Have you check if you have the good column lenght at detector, good make-up flow, leak, capillary adaptor for ECD ect...?
I use Argon / Methane 5% for make-up gas, 60ml/min, 330celcius...
6890 micro ECD is more sensitive than 5890....very more!
Have you check if you have the good column lenght at detector, good make-up flow, leak, capillary adaptor for ECD ect...?
I use Argon / Methane 5% for make-up gas, 60ml/min, 330celcius...
2 posts
Page 1 of 1
Who is online
In total there are 23 users online :: 2 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Georguslat, Semrush [Bot] and 21 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Georguslat, Semrush [Bot] and 21 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science