Electron Capture Detector [May 24, 2004]

Posted: Mon Aug 30, 2004 5:25 pm
by admin
By Ian on Monday, May 24, 2004 - 12:23 am:

We have identified a sample which contains fluorinated and chlorinated hydrocarbons. So far we have determined these compounds using either thermal desorption or headspace as sample introduction devices, followed by GC-MS. We believe an ECD may provide us with a more rapid, sensitive and selective analyses. The problem is that we have no experience with such a detector. I am aware of the detection mechanism of the detector, but if my mind serves me well, I remember reading about a new type of detector that does not operate with a radio active source. Any comments will be appreciated.

-------------------------------------------------------------------------------------------------------
By Anonymous on Wednesday, May 26, 2004 - 08:21 am:

Ian,
Check with Valco for a pulse discharge detector. It can be operated in an electron capture mode and there is no radioactive source used. Web site is www.vici.com I do not work for them, just a very satisfied customer of that detector.
Regards,
Mark

-------------------------------------------------------------------------------------------------------
By Anonymous on Wednesday, May 26, 2004 - 12:06 pm:

Is the PDD as sensitive as a micro ecd of agilent?

-------------------------------------------------------------------------------------------------------
By Anonymous on Thursday, May 27, 2004 - 07:45 am:

May 26 Anon,
Don't know, never used the Agilent detector but Valco might have a comparison. If not I would think they should have sensitivity numbers specs you could use to compare with Agilents numbers. Another nice feature is how fast ther detector recoversm from a large overload. Whereas a "normal" ECD can takes tens of minutes to hours sometimes the PDD will recover in just a few minutes.
Regards,
Mark

-------------------------------------------------------------------------------------------------------
By Ed Chen on Sunday, July 11, 2004 - 12:57 am:

Some fluoro compounds will respond to the ECD while others do not-it could be very selective for some of your compounds

He and Ne separation [May 18, 2004]

Posted: Mon Aug 30, 2004 5:26 pm
by admin
By graciella on Tuesday, May 18, 2004 - 08:14 am:

Please, Does anybody have experience in separating He and Ne with N2 as carrier gas?
Could you advice me which column should work best?

-------------------------------------------------------------------------------------------------------
By Bernd Mischke on Friday, June 11, 2004 - 03:20 am:

A Molsieve 5A can do this job. What kind of column depends on Your GC system.
Important is also the concentration rangeof Your sample.
Contact me.