Advertisement

alternative ferrules for MS interface nut

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
hello,

We use vespel/graphite ferrules (85% - 15%) to connect our capillary column 0.25 mm ID 30 m. to the MSD interface. The MS is an 'old' Agilent 5972. The problem is that we daily have to retighten the nut to avoid leaks. This is very boring, I think it has to do with the ferrule material. I heard there exist a new type of ferrules, consisting of aluminium and vespel. This should be a valuable alternative. Who can give advise? Thanks,

Steven

We also have an "old" 5972 MSD as well some newer ones. We use SGE Sil-tite metal ferrules for the interface to seal well. I have that listed as part #73200.

Thanks,

We also have an "old" 5972 MSD as well some newer ones. We use SGE Sil-tite metal ferrules for the interface to seal well. I have that listed as part #73200.
3 posts Page 1 of 1

Who is online

In total there are 31 users online :: 0 registered, 0 hidden and 31 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 31 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry