ammonium analysis using HPLC
Posted: Tue Mar 06, 2007 11:32 am
by Louise77
Can any one suggest to me any referrence for ammonium analysis using HPLC? Sample will be supernatant from cell extract. Thanks
Louise
Posted: Tue Mar 06, 2007 1:16 pm
by SIELC_Tech
Here is method for ammonia and other simple amines. You need ELSD detector to detect:
http://www.sielc.com/compound_005.html
Contact me if you have questions.
Vlad
Posted: Thu Mar 08, 2007 11:01 am
by Markus Laeubli, Metrohm
Cation chromatography with conductivity detection will be the easiest way to analyse ammonium ions.
Ammonium does not absorb UV. Therefore UV detection will be difficult.
With the respective
column and a
conductivity detector it will be possible to run such sample even on a standard HPLC system.
Posted: Thu Mar 08, 2007 10:16 pm
by XL
A suitable cation-exchange column with suppressed conductivity detection can do the job.
The two links below provide some information of this approach.
http://www1.dionex.com/en-us/columns_ac ... 34323.html
http://www1.dionex.com/en-us/columns_ac ... s2722.html
Ammonium analysis
Posted: Thu Mar 08, 2007 10:25 pm
by maria.rey
The easiest approach will be as mentioned in earlier responses ion chromatography with conductivity detection. Depending on the concentration levels in which other small ions may be present, you will have to make an informed decision on which cation exchange column to use. Supressed conductivity detection will give you better minimum detection levels and more importantly, more options on your eluent concentration, thus allowing the use of cation exchange columns of higher capacity. In case of resolution problems with other closer-eluting ions such as sodium, it will also enable the use of gradient elution.