Separation of Neon from H2 in Helium

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
Hi,
I would like to know which column is the best to use(packed column if possible) and which lenght to separate Neon from H2 in Helium balance.

thank you

dany
On a 6m Carboxen 1000 1/8" OD 60/80 mesh column at 40°C with 30 mL/min carrier of helium . . .

the retention time of Neon is 1.56 min. and is easily resolved from the other gases (except Helium, it elutes about six seconds before Neon) .

Hydrogen elutes at 2.04 min. Oxygen elutes at 7.23 min. Nitrogen elutes at 7.81 min. Carbon Monoxide elutes at 11.17 min.

Methane elutes much later and Carbon Dioxide elutes sometime next week unless you raise the oven temperature. One can elute C2 and even C3 hydrocarbons from this column.

Email me for chromatograms, Deny.

best wishes,

Rodney George
Senior Research and Development Scientist
Gas Separations Research
Supelco
595 North Harrison Road
Bellefonte, PA 16823

814-359-5737 voice
814-359-5459 fax
rodney.george@sial.com
2 posts Page 1 of 1

Who is online

In total there are 22 users online :: 0 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 22 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry