fragmentation in ion source

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
Recently, I try to tune MS (quattro LC) with the synthezied compounds. The problem is that when cone value is low I mainly get sodium adduct, when the cone value is high, I got a lot of fragments. This is just Q1 tuning. I got good {M+H}+ with API 2000 with little fragments. What cause the problem, is it because of Z spray? I use 5mM ammonium acetate as buffer. 50/50 h2o/acn.

We have a Quatro Micro and a 4000 Qtrap. A few compounds form M+Na adducts on the Quattro Micro more than 4000 Qtrap. However, generally I get either M+H ions or M+NH4+ ions at low cone voltage. The M+Na signal tends to increase at higher cone voltages (>50).

For the few compounds that gave primarily M+Na, I tried playing with acid content and every variable I could think of..

However, very little change.

Might look at your tubing and make sure not contaminated with salt from some other work? I guess some sodium could come from your glass reservoirs, but not too likely. We use glass solvent reservoirs all the time.
Sailor
2 posts Page 1 of 1

Who is online

In total there are 9 users online :: 1 registered, 0 hidden and 8 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 8 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry