 - Posts: 357
 - Joined: Wed Aug 03, 2005 1:15 pm
 
						You are seeing the -ve ion equivalent of [2M + H]+, that will often be observed at higher sample loadings.  Depending on the sample loading and structure, the "dimer" signal may be anywhere in the range 5-50% of [M + H]+. 
Regard it as confirmation of MW determined from [M+H]+ or [M-H]-.
JMB
									Regard it as confirmation of MW determined from [M+H]+ or [M-H]-.
JMB
- 
								
 - Posts: 216
 - Joined: Mon Apr 04, 2005 12:26 am
 
						The amount of dimer, 2M-1, depends heavily on the entrance voltage of the skimmer.  For example, look at the ratio of OAc- at m/z 59 to m/z 119 for OAc- . HOAc as a function of the voltage.  Makes a big difference in electrospray..
					
									Sailor
				
			3 posts
							Page 1 of 1
					
	Who is online
				In total there are 15 users online :: 0 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
 
		
Users browsing this forum: No registered users and 15 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 15 guests
Latest Blog Posts from Separation Science
    Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
    
    
    Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
    
    
- Follow us on Twitter: @Sep_Science
 - Follow us on Linkedin: Separation Science
 