Glycoprotein Sepn with HILIC Columns
Posted: Thu Mar 01, 2007 5:35 pm
by mprajesh
Dear Fellow Chromatographers,
I'm trying to develop a method to enrich glycoproteins from non-glycosylated proteins. I'm currently using a HILIC column with ammonium acetate and acetonitrile. The gradient starts with high ACN and then moves to high buffer compositions.
At 70 to 80% ACN starting compositions, the proteins appear to ppt out in the sample! With 50% ACN they are appear ok, but the glycoproteins are retained poorly!
My question: Has anybody used HILIC for separation of proteins? (I know there is literature for peptides).
Has anybody used lectin affinity chromatography columns for glycoproteins? (I have not come across any literature so far!)
Posted: Fri Mar 02, 2007 4:36 am
by Kostas Petritis
Why don't you use an hydrazine resin to enrich the glycoproteins?
Posted: Fri Mar 02, 2007 6:00 pm
by mprajesh
Hydrazine resin? Could you please let me know whether there is any literature about this. Please let me know if this is available commercially.
Posted: Fri Mar 02, 2007 6:48 pm
by Kostas Petritis
Ruidi Aebersold introduced the method some years ago. Below I include an article from him but there are more from his and other groups. The hydrazine resin is available from several vendors (i.e. Bio-rad, Pierce etc) but there is no kit available (i.e. you will have to make your own reagents)... actually have a look around, as I haven't something might have been introduced lately.
Title: Identification and quantification of N-linked glycoproteins using hydrazide chemistry, stable isotope labeling and mass spectrometry
Author(s): Zhang H, Li XJ, Martin DB, Aebersold R
Source: NATURE BIOTECHNOLOGY 21 (6): 660-666 JUN 2003
Posted: Sat Mar 03, 2007 4:46 am
by Uwe Neue
A high percentage of acetonitrile is commonly used for protein precipitation, so I do not think that your HILIC method will have a lot of success. Maybe with another organic solvent like methanol you may have more success.
Kostas: how about boronate resins? Has this ever been used for glycoproteins? Could be even more specific....
Posted: Sat Mar 03, 2007 4:57 am
by SIELC_Tech
mprajesh,
We just introduced new mixed mode HILIC/ion-exchange columns at Pittcon. You can increase amount of water up to 20% compare to traditional HILIC (70% vs. 90%) and this will help you with solubility issue You compounds will retain based on HILIC and cation or anion exchange mechanisms. The phase has both ionizable cation exchange and anion exchange groups. Multiple interactions can be adjusted by pH and buffer concentration:
http://www.sielc.com/Products_Obelisc.html
If you send us sample we can try it in the lab (free of charge)
Vlad