column stability in acetic acid?
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
6 posts
Page 1 of 1
I'm running an analysis once a day at a different location using a Nova-Pak C-18 column and a mobile phase of methanol/water (30:60) with one percent acetic acid. There isn't time to flush the mobile phase each day and replace it with methanol/water. Is it safe to just shut off the flow when I'm not analyzing and leave the column in the acetic acid mobile phase?
-
- Posts: 95
- Joined: Mon Feb 12, 2007 10:05 pm
I run almost the same mobile phase and have instruments idle for a couple of days every few weeks. I haven't seen any detrimental effects. As long as the sample components have eluted completely, you probably don't need to worry. I replace my columns every 6 months as a preventative measure but that's after 7000+ samples.
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
We oftentimes leave mobile phase with 0.4% acetic acid in the aqueous phase in the columns for weeks or longer. No issues.
-
- Posts: 397
- Joined: Fri Sep 22, 2006 7:36 am
When I am using the same mobile phase every day, at the end of the day I flush the column with pure solvents. For example, if i'm using metanol/water with one percent acetic acid, at the end of the day I would flush the column with the same ratio of metanol and water (but pure,without acetic acid).
I think that kind of column care can prolong the column lifetime, although acetic acid does not make problems at the moment.
Best regards
Zoran
I think that kind of column care can prolong the column lifetime, although acetic acid does not make problems at the moment.
Best regards
Zoran
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
We only flush if we have dissolved solid modifiers in the mobile phase. That's one reason we only use such solid modifiers 10% of the time (only when really needed).
-
- Posts: 442
- Joined: Mon Jun 13, 2005 7:03 pm
Thanks for the replies. You've set my mind at ease and saved me a lot of extra work. Usually I do flush out the columns after the analysis is complete but in this case it would be very inconvenient to do so. Isn't this forum great?

6 posts
Page 1 of 1
Who is online
In total there are 20 users online :: 0 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 20 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 20 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science