2 TSKgel superSW3000 columns in series + guard column
Isocratic elution, 0.35 ml/min, PBS, pH 7.4
30 min run for each sample
10 mg/ml mAb, 10 ul inj
stability samples, so some are clean mAb with little aggregation, some are stressed, with 20% aggregration. all samples were filtered through 0.45um before inj.
reference standards were injected every 10 samples in any sequence, from same vial. a typical sequence runs for 20 hours. samples were kept at 4 degrees C in autosampler.
Observation:
In a sequence of 20 to 40 samples, back pressure started normal, after 6 or 10 or 15 sampels, back pressure started going downing, sometimes it could drop by 20% in a few samples. Then back pressure went back to normal level after being lower for several samples. I believe that when the pressure was low, flow rate was low because the PA of those samples were abnormally higher than usual.
This has been reproducible, observed in many sequences. The HPLC is an Agilent 1100, technician came in a couple times and didn't see anything wrong with the instrument. Tosoh didn;t know what happened.
Once I tried injected 20 injections of water, similar thing happened.