the absorbance should be less (elute comes out slower) than that at normal flow rate
I see where you're coming from (but trust me, you've got it wrong).
The analyte comes out more slowly, but so does the solvent; the concentration therefore stays the same. Peak area is essentially the product of concentration and residence time in the flow cell.
Imagine the following experiment: Inject a sample under your normal conditions and wait until one of your peaks starts to elute. At the very top of the peak, instantly stop the flow (I know you can't
really instantly stop the flow, that's why it's an imaginary experiment). What will happen to the signal?
*
*
*
*
*
*
*
pause to let the suspense build!
*
*
*
*
*
*
If you were using a mass-sensitive detector, the signal would drop back down to zero (think of a flame ionization detector in GC: once you burn all the sample, it's gone).
In a concentration-sensitive detector (like UV), the sample is still in the flow cell, still absorbing light. If you waited 5 minutes and then instantly turned the flow back on, the signal would start dropping again; you would end up with a huge, flat-topped peak; something like this (with apologies for the crude artwork!):
The area in this case would bear no relationship to the amount of sample injected; by changing the duration of the stop-flow interval, you could get any area you wanted.