What is ODS reversed-phase column. The column i am trying to use is only reversed-phase column C18
ODS = "OctaDecylSilyl" = C18
However, there are over 500 different C18 columns on the market and they vary tremendously in their chemical characteristics.
Can i adapt the method from ODS reversed-phase column to the column i have.
Most separations can be done on most C18 columns, so the answer is "probably yes", but the key word in your question is "adapt". You will probably have to adjust some of the other parameters (organic solvent, organic:water ratio, pH, temperature, buffer concentration, etc.)