The process is a mammalian cell culture fermentation followed by primary clarification and then a chromatographic purification chain.
Produces a wel-characterised glycosylated protein product
Is reproducible and controlled
Is developed from laboratory validated pilot.
The chromatography operating cycle inludes a gradient elution step
The validated allowable ranges of operation have been separately determined for : column capacity; load rate; allowable number of reuses and hold times.
So the intended process change is:
to install a larger fermenter and primary clarification chain but use the same column series.
The major issue remains:-
Does the change to larger fermenters change the product composition in any way that might compromise the ability of the chromatography to purify the product?.
Because changing fermenter size is a major step ( refer my earlier post ) and it seems that you are not using defined media, most regulators would expect to see clear experimental evidence that the product profile and gunk requiring separation have not changed due to the scaleup.
Fermenters produce gunk, and the quality and quantity of the gunk varies according to operational conditions and imputs, including media.
The fact that the laboratory process was validated is only relevent if the scale up process has also been validated to the new size. If you can show that the increased size of the fermenters has no effect on the output, then the chromatography qualification become trivial. If you can't, then you probably are facing a a fairly significant process requalification, with the associated analytical nausea.
All of these aspects should have been identified in the initial risk assessment performed by the quality group.
As I noted above, I'm not an expert on fermentations, but the first question I would ask is, " How do you know that the product and gunk are still identical to the previous material going into the chromatography?.
If you can show it is, then the risk assessment should have indicated that any qualifications of the chromatography, due to the more intense usage, should be minor. The fact that you are using chromatography to purify the product indicates that the unwanted gunk has also to be unchanged to ensure the chromatography is not compromised.
As I stated earlier, fermentation is not my field, there are consultants that specialise in the regulaory control of fermentation-derived products, and they would be the best people to ask.
Bruce Hamilton