We observed with a mAb, both in SEC and CEX, PA increased almost linearly when reducing flow rate.
Is that unusual?
Thanks!
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
I really need to add this one to the FAQ ("Frequently Asked Questions") section, because it comes up with fair regularity.
Assuming a concentration-sensitive detector (such as UV), the peak area is obtained by reading the signal (absorbance) at regular time intervals, and adding up all the readings from the beginning to the end of the peak.
To a first approximation, if you cut the flow rate in half, the sample spends twice as much time in the flow cell, you get twice as many readings, and the area becomes twice as large. What you are seeing is perfectly normal.
the absorbance should be less (elute comes out slower) than that at normal flow rate