Dear all,
For choosing the concentration of internal standard, we always/usually choose the concentration corresponding to the CCS-7 of aqeous solution. what is the reason of choosing specially cc7 when we use cc1 to cc8. if i choose some other level will it affect my result.
with regard
Donal

Please do not bombard us with incomprehensible algorithms - what is CCS-7 ?

Peter
Peter Apps

Please do not bombard us with incomprehensible algorithms - what is CCS-7 ?

Peter
ccs7 means the 7th standard in the calibration curve for linearity.
with regard
Donal
"if i choose some other level will it affect my result."

Yes.

Will is significantly change your result? Depends.

It would probably help to have some knowledge of what system you are running LCMS?
Are you talking about mass accuracy? or quantification of area / intensity?

For mass accuracy I find best results are obtained with an internal standard of similar intensity to the peak being analysed. If the intensity is close to your cc1 then your standard should be too. Wouldn't it be nice to know this beforehand! Generally a less intense standard compared to a more intense analyte works a bit better than vice versa.
"if i choose some other level will it affect my result."

Yes.

Will is significantly change your result? Depends.

It would probably help to have some knowledge of what system you are running LCMS?
Are you talking about mass accuracy? or quantification of area / intensity?

For mass accuracy I find best results are obtained with an internal standard of similar intensity to the peak being analysed. If the intensity is close to your cc1 then your standard should be too. Wouldn't it be nice to know this beforehand! Generally a less intense standard compared to a more intense analyte works a bit better than vice versa.
actually i an using sciex tripple quad with analyst software and i am talking about quantitatiopn of area not mass. is there any reference so that i can go through regarding the statistic part. to be very frank i am not statiscian but just want to know the reason. just simply following the seniors does not seems good. thank you.
with regard
Donal
5 posts Page 1 of 1

Who is online

In total there are 46 users online :: 0 registered, 0 hidden and 46 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 46 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry