very low peak height and negative peak
Posted: Wed Feb 21, 2007 4:31 pm
Hi, all
I have a question quite urgent. I'm a new beginner in HPLC with a old machine, that is definitely worse than a new machine when you get no information about it. Before the analysis, I tried to check the status of it. Although I faced some problems with that machine: noisy in the baseline, high pressure, it all comes from the old column. After I purchase a new column, it seems to be OK. The column is Lichrosorb Si 60, 4.6 mm ID, 125mm, normal phase column, the guard column are coated with same material. My compound is Canthaxanthin (carotenoid). The new problem comes, the parameters I set as below:
Flow rate: 1ml/min
UV/Vis: 470nm
Mobile phase: n-hexan/acetone 86:14, V/V
Test solution: 1mg Canthaxanthin in 100ml Mobile phase.
The problem is I get very low peak height (0,0002 or 0,0003 volts)after 2 min, and then there is no peak any more.
I took off the column and tried to inject directly into the UV detector, the peak appeared and the peak height is around 0.2 volts.It seems my detector is ok.
I guess the problem maybe the concentration of test solution is too high. but I'm not so sure...
And today I tried to repeat the experiment at 0.5 ml/min flow rate, the only peak I mentioned above becomes more sharp and seems to be the peak of my compound, but still very low peak height, (Apr.0,006 volts, 10 times more than before),is my column has some problem? I hope not, it's a new one. and after this peak, there is a small negative peak, how can I do with that? Actually, I think the peak height even it is very low, but still make sense for my work, at least I get a peak, but negative peak, what can I do?
Thank you all
I have a question quite urgent. I'm a new beginner in HPLC with a old machine, that is definitely worse than a new machine when you get no information about it. Before the analysis, I tried to check the status of it. Although I faced some problems with that machine: noisy in the baseline, high pressure, it all comes from the old column. After I purchase a new column, it seems to be OK. The column is Lichrosorb Si 60, 4.6 mm ID, 125mm, normal phase column, the guard column are coated with same material. My compound is Canthaxanthin (carotenoid). The new problem comes, the parameters I set as below:
Flow rate: 1ml/min
UV/Vis: 470nm
Mobile phase: n-hexan/acetone 86:14, V/V
Test solution: 1mg Canthaxanthin in 100ml Mobile phase.
The problem is I get very low peak height (0,0002 or 0,0003 volts)after 2 min, and then there is no peak any more.
I took off the column and tried to inject directly into the UV detector, the peak appeared and the peak height is around 0.2 volts.It seems my detector is ok.
I guess the problem maybe the concentration of test solution is too high. but I'm not so sure...
And today I tried to repeat the experiment at 0.5 ml/min flow rate, the only peak I mentioned above becomes more sharp and seems to be the peak of my compound, but still very low peak height, (Apr.0,006 volts, 10 times more than before),is my column has some problem? I hope not, it's a new one. and after this peak, there is a small negative peak, how can I do with that? Actually, I think the peak height even it is very low, but still make sense for my work, at least I get a peak, but negative peak, what can I do?
Thank you all