I'm posting to share an experience with a particular British column distributor to hoping to find whether I am receiving a typical degree of support. The situation relates to us attempting to get what we feel is a faulty column replaced as quickly as possible.
- Our US client develops an HPLC method, purchases an second equivalent column of different stat. phase batch and confirms the method 'works'.
- Client sends the newer column to us for method pre-val and utimately validation. We purchase a second equivalent column from the UK distributor. It's of different stat. phase batch.
- Chromatography on the UK column is immediately very poor; main peak elutes late, is split and tailing. US column on our system looks good; profile consistent with data supplied by the client.
- We contact and explain the problem to the UK distributor. They want our column back to test before drawing a conclusion. We have not run the test mix depicted on the column certificate.
Is it others' experience that this is the way of things? Are we reliant on the distributor confirming our poor column performance? We have here, and have also read, examples of confirming system suitability by just running the method. Do you think the delay would of been avoided if we had run the manufacturer test mix?