Re! Limit of detection for Bromo Benzene by HS-GC analysis
Posted: Thu Feb 15, 2007 6:48 am
by kodumuru
I am just looking for possible " lower Limit of detection for Bromo Benzene by Headspace GC"analysis along with the GC parameters.
Thanks[/b]
Posted: Thu Feb 15, 2007 10:34 am
by GOM
Hi,
I'm sorry to sound pedantic but there are a lot of "it depends".
For example;
Static or dynamic headspace?
What is the sample matrix?
How much sample have you got?
What is your detector?
As an example, and without having done it, I would expect to be able to see 100ppb or lower using SPME/GCMS-SIM on 1g sample heated to 50°C, and sampled for 20mins using a standard addition for quantitation, analysed on a wax column starting at 40°C.
Regards,
Ralph
Re! Bromo Benzene HS-GC analysis
Posted: Thu Feb 15, 2007 12:16 pm
by kodumuru
The system config as floows.,
Dynamic headspace (G1888 - Agilent)
FID detector (Agilent GC 6890n )
Sample concentration about 100 mg/ml (expected bromo benzene level is about 0.5 ppm )
Sample matrix contains Methanol , Acetone and Bromo Benzene which are used in the process.
We are interested to carry the analysis on GC Capillary columns.
Thanks
Just a warning
Posted: Thu Feb 15, 2007 1:27 pm
by chromatographer1
Be aware that it is possible to de-brominate your analyte. Given active sites in your headspace system, reactions can take place and you can easily loose low levels of the bromobenzene.
good luck in your analysis.
Rod