- Posts: 63
- Joined: Tue Oct 24, 2006 10:16 pm
1 - I was told by someone to never let a GC column go "dry"; that is to always have a flow of the carrier gas (in my case, helium) unto the column since otherwise, it would cause damage to the column (HP-1).
But isn't this only if the oven is on? Doesn't oxygen attack the silica only when it is heated? If the oven is off, the column should be safe whether there's a flow of helium or not, no?
2 - Overnight, we put our GC's in a "sleep mode" that was set by the Agilent rep. It sets the inlet mode to splitless, with a column flow of 3.0 mL/min and the GasSaver option is on and set to 20 mL/min at 2.00 mins - therefore giving a total flow of 25.8 mL/min. The "purge Flow to Split Vent" is set to 10.0 mL/min at 1.00 min so if we disable the GasSaver option, the total flow falls to 15.9 mL/min.
My question is, are such high flows necessary? We shut down the oven, detector and inlet heating at night. In order to save gas, wouldn't it be more logical to bring the "purge flow to split vent" to 0 mL/min and set the column flow to something like 1.5 mL/min...? Or is this a bad idea...?
Thanks,
Roxanne.