Failing Initial Calibration on HP 5890

Discussions about GC and other "gas phase" separation techniques.

5 posts Page 1 of 1
Hello people,

I run PCBs on a HP 5890 ECD GC. For the past month or so I have had to run an initial calibration on my instrument almost every other day.

It can't be contamination from my samples because I am only running MDLs on that instrument.

I completely cleaned out my injection port, put in a new liner, septum, gold seal, syringe and replaced the weldman and gas lines but still had the same problem.

Next I changed the split ratio from 5:1 to 12:1 and changed my injection volume from 1ml to 2ml, but now I can't my initial calibration to pass.

I am almost out of ideas .... how should I proceed ?

Somebody help me !!!!

Thanks..

Hi!
First,why do you have change or clean injector, liner, ect....?what was your problem for change this item?

I changed the liner and cleaned out the instrument to rule out the possibility of contamination from the instrument.

Hi!
Your initial calibration don't pass.....why?...give your symptoms...area count, sensitivity, tailing,no or lost of peak??

How old/used are your detectors? Have you considered the possibility that they may need cleaning? That is a common problem with a local
laboratory I work with. Their 5890 ECD's will show slightly elevated standing currents and not be able to hold calibration.
5 posts Page 1 of 1

Who is online

In total there are 4 users online :: 3 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Peppels, satya5 and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry