Method validation accuracy
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
4 posts
Page 1 of 1
It is acceptable to determine method accuracy by comparing the results to the historical mean obtained by the method under validation rather than against a known reference standard?
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
That question is a bit too broad to allow a simple answer.
Is this an existing method that you run regularly or are you re-validating an old method or are you validating a new method?
Is this a "regulatory" method (i.e., one that will have to meet FDA (or equivalent) scrutiny?
What are you using as calibrators (standards) on a day-to-day basis?
Is this an existing method that you run regularly or are you re-validating an old method or are you validating a new method?
Is this a "regulatory" method (i.e., one that will have to meet FDA (or equivalent) scrutiny?
What are you using as calibrators (standards) on a day-to-day basis?
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 5
- Joined: Mon Sep 13, 2004 5:25 pm
This is a retrospective validation for a protein (P) used for regulatory compliance. Accuracy is determined by using known conc. of P and calculating the conc. against a standard curve of BSA. However, the known conc. of P is determined based on the historical assay results. There has been no independent verification of P’s value. Can accuracy be properly assessed is the known value of P has not been independently verified?
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
Well that helps provide the basis for a short answer:
"No".
As I read ICH / FDA, you either a known standard or comparison with an independent (known to be accurate) technique.
"No".
As I read ICH / FDA, you either a known standard or comparison with an independent (known to be accurate) technique.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
4 posts
Page 1 of 1
Who is online
In total there are 13 users online :: 1 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Semrush [Bot] and 12 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Semrush [Bot] and 12 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science