Hi WK
The air and nitrogen are used for the modulator hot and cold jets respectively. Running GCxGC all day you can get through a cylinder of each in three to four days - it is difficult to be exact because we run other GCs from the same cylinders. If you have any FIDs in the lab you need to run their air and nitrogen from a separate regulator from the modulator or else you get baseline deviations due to the modulator switching the flows on and off.
I do not know if the Leco modulator can work with liquid CO2, certainly the set up we have can only use liquid nitrogen. The liquid nitrogen is used to chill the nitrogen from a cylinder that blows into the modulator.
The raw data files are multiple megabytes, even with the top end computers that come with the instrument, number crunching takes several tens of minutes, and if you set a peak detection threshhold too low it can go on for hours.
Target pesticides in a GC compatible matrix like an essential oil is one area where GCxGC TOFMS does have advantages - Leco would not have published it otherwise of course. As with all manufacturer's application notes you need to be aware that the notes are deigned to sell instruments - check out the concentrations of analytes that produced the peaks that you can see e.g. fig 6 is with 1000 ppb, while the repeatability data in Table 1 is at 25 and 50 ppb, when the target peak is most likely invisible to the human eye and has to be dug out from the background by the deconvolution algorithm. I am an old fashioned kind of chromatographer - I like to see the peaks myself !.
If you are interested in pesticides at low ppb, spike some oils at the appropriate level and see what Leco can do with them in terms of detection, quantitation and repeatability - that is really the only way to know if the machine can do the job that you want it to. You could also give the same samples to other vendors to see what their instruments make of it - Agilent also has deconvolution, but not GCxGC as far as I know, maybe accurate mass SIM would pull the pesticides out of the terpene background. A heartcutting approach with conventional MS might also work, you then maximise the GC separation rather than the MS selectivity. There are several ways to skin this particular cat.
Peter