transfer line GCHS
Posted: Mon Jan 29, 2007 3:57 pm
by raju
Hi all. I have two questions.
1. I developed a gc-hs method in thermo finigan GC-HS that does not have transfer line. when transfered to a Perkin elmer GC-HS the peaks do not resolve, What needs to be done when you transfer such methods.
2. Is transfer line GC-HS a must to follow USP <467>.
Posted: Mon Jan 29, 2007 5:44 pm
by chromatographer1
The Perkin Elmer 'transfer line' can have the column inserted all the way to the needle if I remember correctly. My HS-40 certainly could, although I used a fused silica deactivated piece of SS tubing and a butt connector in the GC oven to connect the two.
Is this similar to your Thermo HS analyzer hardware? Or does it use a syringe?
Otherwise, there must be a transfer line of some sort. How do you connect your sample vial to pass your gas sample to your column?
best wishes,
Posted: Tue Jan 30, 2007 6:35 am
by Peter Apps
Hi Raju
Is the loss of resolution becuase the peaks are broader, or because their retention times have changed ?
u can try this
Posted: Mon Feb 12, 2007 6:12 am
by dhruvil
Hi
Raju
This is dhruvil. Oftern this problem is observed when u transfer a method which is developed on Shimadzu or some other make GC onto Perkin Transfer line GC.
U can adjust this parameters or just check what values u have adjusted on yr GC (perkin)
Yr syringe temp, conditioning temp. and the conditioning time u can keep the same as yr original method. But u cannt put 01 ml sample injection volume with Perkin transfer line GC. That u keep 0.2 minute, Pressurization time as 0.3 min.
i.e yr injection time will be 0.2 min
Pressurization time will be 0.3 min.
Yr keep yr signal attanuation -2 or -4. (optimise it for yr method.)
U try yr method with these settings. i think u should get perfect resolution and peak shapes and everything.
This problem is faced with Perkin because it has got very high sensitivity compared to other GCs. ok
ok then try this and let me know.
all the best :D