I am in the process of chasing down leaks on my HP5972 mass analyzer. The analyzer seats on top of an oval-shaped gasket, which is seated on top of the main chamber. The system pumps down fine, but air/water checks routinely come up with high 28/69 ratios (last one was 49%). When I run a scan and spray "dust-off" (office depot, HFC134-A, m/e~33, 51, 69, 83) signal rapidly appears when I spray the area around the gasket, near where the transfer line comes in. (This is after having vented the system, cleaned the gasket and surrounding area with IPA, and reseated it). When I put the cover back on the analyzer and spray other locations (Xfer line, Xfer line-column connection) no signal appears. (Leak tests were done with the injector port blanked off, so these leaks are discounted). Any ideas? Might the gasket be upside down?
Rich