Anybody in here run Methods 504.1 or 8011?
Posted: Mon Jan 22, 2007 7:20 pm
by chemwipe
Hello People-
I need to start running my EDB/DBCP samples with surrogate, and I'm not sure what to use. I haven't found anything online - just that a surrogate needs to be run. I've tried 1,3-dichlorobenzene, but didn't get great recovery consistency on my samples (anywhere from 40% to 90%). The calibration for this surrogate worked okay though.
Any suggestions?
CW
Posted: Tue Jan 23, 2007 6:15 am
by chemwipe
Anybody?
504.1
Posted: Fri Jan 26, 2007 8:34 pm
by vballchemist
We run 504.1 with TCMX and DCB. The same used for pest/pcb's. They both elute later than the targets analytes and are very stable.
Re: 504.1
Posted: Mon Jan 29, 2007 8:34 pm
by chemwipe
What concentrations for TCMX and DCB seem to work well?
My calibration points for 504.1 are
0.20 ppb
0.15 ppb
0.10 ppb
0.04 ppb
0.02 ppb
I calibrated with 1,3-Dichlorobenzene as my surrogate at the above levels and it seemed to work okay. I'm not against trying TCMX and DCB.
What kind of columns are you using? I use Restek CLPesticides (I and II) columns.
Thanks.
CW
Posted: Mon Jan 29, 2007 9:01 pm
by vballchemist
We use the same columns, 0.53mm id. Our curve is:
0.2ppb
0.1ppb
0.05ppb
0.02ppb
0.01ppb
The surrogate compounds are 4ppb.
Rarely any problems with recovery.
Posted: Mon Jan 29, 2007 9:08 pm
by chemwipe
What levels do you calibrate for the surrogates?
Posted: Mon Jan 29, 2007 9:28 pm
by vballchemist
We calibrate using the average response of 4ppb surrogates in all curve points.
Posted: Mon Jan 29, 2007 9:41 pm
by chemwipe
Cool. Thanks!