Grace,
I hope my answer does not put you off on your first visit but more information is needed, such as why do you want to analyse these mixtures, what is the application, what concentration, how are you going to introduce the sample on to the column? Do you have to use GC. Can't you use direct-reading instruments?
Without a direct-reading instrument corrosive gases like NH3 SO2 HCl are almost always a job for ion chromatography, not gas chromatography. Sorry to be negative but for GC you would need a packed column with a custom made porous polymer. Others here may have suggestions, but think about how corrosive gases are going to affect your TC detector.
Mass spectrometry? Yes it could be done with a piece of kit called a process mass spectrometer, no gas chromatography needed. It would be custom solution and expensive.
Without knowing more about your application what I have just said may be of no use to you, sorry.
Mike,
Thank you very much for your reply!
The gas concentrations are about 10-100ppm, introduce samples using a gastight syringe. The purpose of these analysis is just validating a gas supply system.
From our current experimental condition, I have to use GC/TCD FPD and a capillary column even these corrosive gases may damage the instrument parts.
But I'd like to know more about the direct-reading instruments and process MS, do you have more information about these like websites, documents?
Best,
Grace