dhruvil » Mon Feb 12, 2007 6:12 am
Hi
Raju
This is dhruvil. Oftern this problem is observed when u transfer a method which is developed on Shimadzu or some other make GC onto Perkin Transfer line GC.
U can adjust this parameters or just check what values u have adjusted on yr GC (perkin)
Yr syringe temp, conditioning temp. and the conditioning time u can keep the same as yr original method. But u cannt put 01 ml sample injection volume with Perkin transfer line GC. That u keep 0.2 minute, Pressurization time as 0.3 min.
i.e yr injection time will be 0.2 min
Pressurization time will be 0.3 min.
Yr keep yr signal attanuation -2 or -4. (optimise it for yr method.)
U try yr method with these settings. i think u should get perfect resolution and peak shapes and everything.
This problem is faced with Perkin because it has got very high sensitivity compared to other GCs. ok
ok then try this and let me know.
all the best
Regards,
Dhruvil