My column: Thermo Hypercarb 150mm X 1mm; 5um
I have the technical data sheet attached together with the column that I've purchased and the column back pressure was noted at 1100psi. What does that mean? Wasn't it suppose to be the maximun allowable pressure for the column?
What does back wash means? Is it using the column on the reverse side?
When I tried to recondition the column by flowing the solvents through the column with the HPLC system, I cant get the pressure to get lower; in fact, the pressure keeps on going up even though the end of the column was not attached onto the LCMS-MS when I try to increase the flow rate through the column.
Does that mean the column has choked or something or the frit has clogged? Do advise recommended actions to be taken.
Solvents tried for flushing the column: ACN:iPA (1:3) and THF (100%) (I did not try to flush the column in reverse order)
If you have the technical data sheet for me to note the maximum allowable pressure for the column as well as back washing or running the column backwards; that would be really helpful for me. Do attach to me too.
P/S: BTW, just curious; since reverse use of the column can be done, why there is an indication note that stated which side the flow should be?
Do let me know ASAP. Thanks a lot!!!














 
																							 
					