Calculate concentration
Posted: Sun Nov 26, 2017 4:40 pm
by snhafizai
Does anyone know how calculate the concentration from a gc/ms chromatogram which is having retention time and peak area only? I don't have any internal standard.Thank you
Re: Calculate concentration
Posted: Sun Nov 26, 2017 8:30 pm
by tom jupille
To *grossly* oversimplify: you run "calibrator" standards at known concentrations (ideally at 5 levels) and plot peak area vs. concentration. If all goes well, you will see a linear relationship; the slope of the best-fit straight line is the response factor (RF). Now run your sample(s); divide the peak area by the response factor to determine concentration.
Posted: Tue Nov 28, 2017 1:32 am
by snhafizai
i don't have 'calibrator' standard of known concentration. i only have peak area and retention time.
Posted: Tue Nov 28, 2017 3:13 am
by HPLC chemist
Then your screwed. You must have an pure internal standard for calibration since the detector response/mass can vary radically between molecules.
Posted: Tue Nov 28, 2017 6:42 am
by Peter Apps
Depending what accuracy you need you might be able to calibrate using a similar compound. What compound is it that you want to quantify ?
The more you tell us the more we can help you.
Posted: Wed Nov 29, 2017 4:23 am
by snhafizai
Hi, i want to quantify the dibenzothiophene (DBT). the initial concentration was assumed 500ppm but i need to calculate the actual concentration of the initial sample in order to get concentration of DBT after the adsorption process. thank you
Posted: Wed Nov 29, 2017 5:58 am
by Peter Apps
Dibenzothiophene is readily available at high purity in small quantities - just buy a standard. Sigma-Aldrich have it, or probably others.
Posted: Wed Nov 29, 2017 1:46 pm
by snhafizai
Is there any other option to calculate concentration since i already finish doing the lab works? Thank you.
Posted: Wed Nov 29, 2017 2:17 pm
by rb6banjo
No. You must calibrate your instrument with a known concentration of the analyte before you run your analysis to determine an unknown. As Peter mentioned, you can use a similar material if you want to get close but you won't know for sure, exactly, unless you use the authentic material to calibrate.
Posted: Thu Nov 30, 2017 7:43 am
by Peter Apps
No, you should have run the calibration when you had the chance I'm afraid. The best you can do now is to compare between samples, e.g. before and after they have been through the reactor (assuming that you ran both before and after samples).
Posted: Thu Nov 30, 2017 7:56 am
by snhafizai
I already compare the result before and after they go through the reactor. However the results vary and i think the initial concentration that i' ve prepared is above 500 ppm.
Posted: Thu Nov 30, 2017 2:02 pm
by Peter Apps
Measuring the concentration of the feed will not make any difference to the variability of the results.
