Flow rate to be kept if not mentioned in USP

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

8 posts Page 1 of 1
Dear all chromatographers and regulatory people,

Can you help me with the query as mentioned below:

Query: In one of the product monograph in USP, Assay method (HPLC) does not mention the flow rate to be kept during analysis. At what flow rate the analysis should be performed. Is it mentioned in USP that if the flow rate is not mentioned we have to run the chromatography at 1.0 ml/min.

Pl reply.

Regards,
Surabhi

If flow is not specified, then it is not specified. You may adjust as necessary to meet any retention time specifications.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

Like many of us who actually work in the lab, you've found that USP is not as thorough in some areas as management or QA thinks.

Hi Surabhi

As mentioned above, it’s obviously up to you to specify the flow rate.
So if the column is specified I’d go for the optimal flow rate in order to achieve the best performance. F. ex. if the column diameter is, let’s say, 4.6 mm and the particle size is 5 µm, I’d set the flow rate to 1 mL/min.

Best Regards
Learn Innovate and Share

Dancho Dikov

To : Surbhi

I understand from your problem that in USP monograph , Retention Time of product or Relative Retention Time of the product is also not mentioned .

Hence You may keep 1 ml / min. Flow rate if it is not mentioned in USP monograph of your desire product for 250 lengthed column & inject your product & Standard 6 times & check the repeatability . See that your desire peak R.T. should be more than dead volume of the column too .


Hope this will solve your problem.
SUNIL PANDYA

Wouldn´t one be inclined to find the optimum flow rate if it was not specified?

Wouldn´t one be inclined to find the optimum flow rate if it was not specified?
Depends on your definition of "optimum" (my personal choice would be "the highest flow rate which gives me acceptable results and doesn't over-stress the system").
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

That´s just fine also, I was thinking more of optimiziing resolution. As long as you do something (like using gray substance) and then state somewhere what you did and why.
8 posts Page 1 of 1

Who is online

In total there are 23 users online :: 1 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot] and 22 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry