My problem is when a gradient is run, it doesn't go to the next step.Just doing the first line of gradient in Agilent 1200 HPLC. Any suggestions? Could this be MCGV?
-
- Posts: 320
- Joined: Tue Jul 12, 2016 4:47 pm
Double check your run time in your sample set, and make sure its not set to lower than the the time required to make the first switch.
Also, make sure the gradient curve in the second line is not set to 11- if so, no change will take place until the end time of the second curve event.
Also, make sure the gradient curve in the second line is not set to 11- if so, no change will take place until the end time of the second curve event.
2 posts
Page 1 of 1
Who is online
In total there are 31 users online :: 1 registered, 0 hidden and 30 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Amazon [Bot] and 30 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Amazon [Bot] and 30 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science