The unit:
P.E. Clarus 500 gc-ms, DB 5 column, UHP He carrier 1 ml/min, E.I. Detector
The problem:
Changed the column after prolonged downtime. Pumped overnight. Good vacuum. HUGE Nitrogen tune peak. LEAKING?
Re-plumbed the column (new ferrules) , Changed to new colum, new ferrules, changed He gas tanks, changed He gas LOT #'s, Squited Freon ALL OVER the system (no freon masses), re-tuned.... slowly loosing control... about to pick up hammer.
after 3 days
m/z
4 = 39%
18 =17%
28= 100%
32= 19%
Help!