Posted: Wed Nov 22, 2017 2:42 pm
by Bminer
Hello,

I have a 7673 autosampler that I am trying to connect to a 5890 Series II GC via a windows XP running Chemstation A.10.02. The computer system and autosampler were previously running a 5890 Series II Plus which has since been decomissioned. The Series II communicates with the PC just fine, but I get an "INJ 1 not configured error" when loading up Chemstation. Is there some address I need to configure somewhere? Or how do I tell if I am running the right hardware/firmware for the autosampler?

Re: 7673 Autosampler communicate w/5890 II

Posted: Wed Nov 22, 2017 3:56 pm
by dblux_
I have a 7673 autosampler that I am trying to connect to a 5890 Series II GC via a windows XP running Chemstation A.10.02. ...
According to ChemStation compatibility matrix 7673A (18594A) is not compatible with ChemStation A.10.02 but 7673B (18594B) and 7673C (G1512A) is compatible.

Re: 7673 Autosampler communicate w/5890 II

Posted: Wed Nov 22, 2017 4:10 pm
by Bminer
Thank you for your reply. My apologies for not specifying - it is a 7673 B autosampler. It has previously worked with this version of chemstation -- I am wondering if there is something different about the 5890 II vs 5890 II Plus that is preventing communication.

Re: 7673 Autosampler communicate w/5890 II

Posted: Wed Nov 22, 2017 5:55 pm
by dblux_
Thank you for your reply. My apologies for not specifying - it is a 7673 B autosampler. It has previously worked with this version of chemstation -- I am wondering if there is something different about the 5890 II vs 5890 II Plus that is preventing communication.
Your both GC and ALS are compatible with ChemStation A.10.02.
http://cn.agilent.com/cs/library/suppor ... a10674.pdf

There are some demands on firmware versions, check with the manual:
https://www.agilent.com/cs/library/user ... enance.pdf

Hope you will make it working.