				[req] Condition for C18 column for Vitamin C Tablets, thanks
				Posted: Thu Jan 04, 2007 12:37 pm
				by kingsemma
				Can anyone tell me the condition and mobile phase for Vit C Tablets using C18 column (Waters Symmetry) and PDA detector?
Thank you very much!
			 
			
					
				[req] Condition for C18 column for Vitamin C Tablets, thanks
				Posted: Thu Jan 04, 2007 2:51 pm
				by skunked_once
				Have you tried a literature search or contacting the vendors for application notes?
			 
			
					
				
				Posted: Thu Jan 04, 2007 5:38 pm
				by Mark Tracy
				Vitamin C is so hydrophilic that it elutes in plain phosphate buffer at pH 3.0 +/- 0.5. The optimum pH depends on the sample matrix and interferences that you want to avoid. The Symmetry column is probably not the best option; a polar-embedded column is the better option. UV detection at 246; the spectrum is rather distinctive. Do you need to distinguish iso-ascorbic? If so, you need a long column.
			 
			
					
				
				Posted: Thu Jan 04, 2007 5:53 pm
				by SIELC_Tech
				Here is a nice alternative using mixed mode approach:
http://www.sielc.com/compound_149.html
You can achieve reasonable retention (5-15 min on 150 mm column) by playing with mobile phase pH and ion strength.
Contact me if you have questions
Vlad
 
			
					
				
				Posted: Fri Jan 05, 2007 6:31 pm
				by Bryan Evans
				If you want to analyze ascorbic acid without adding ion pairing reagents, 
you can try using a silica column.  This separation was done on
our Unison UK-Silica column:
http://www.imtakt.com/TecInfo/TI154E.pdf