Headspace 6890 and 7697 - no trigger to acquire

Posted: Mon Nov 20, 2017 9:53 am
by ydna1977
Sorry -I put this in wrong subject

Hi ,

Headspace 7697A carries out equilibration , pressurization and injection down transfer line
HP6890 then starts oven program (lights on HP6890 turn on)
Empower 3 sits on status 'Preparing - Sample for injection'
No trigger to collect and acquire data
Lace boxes all working and pinging (IP addresses ok) and you can monitor baseline and detector output ok
Any thoughts ? Thanks

Re: Headspace 6890 and 7697 - no trigger to acquire

Posted: Thu Nov 30, 2017 2:55 pm
by ydna1977
Hi ,

Still no forward with this

When the 7697A headspace injects the status time on GC 6890 immediately starts counting upwards (Time = 0.xyz) so injection is recognised from HS to GC
Empower 3 remains on 'waiting on injection'

Any thoughts much appreciated

Re: Headspace 6890 and 7697 - no trigger to acquire

Posted: Fri Dec 01, 2017 7:10 am
by ydna1977
Hi ,

I did notice configuration on back ALS injector

What should this be set at ? Does this come into play with remote control ?
Current options
No Headspace
Headspace 7697A (12 Vial)
Headspace 7697A (111Vial)

But when I select Headspace 7697A (111Vial) the GC out put tries to communicate with Empower and counts up
Downloading method info from Empower to GC and HS not a problem