Page 1 of 2
Headspace 7697A with 7890 - No trigger to acquire
Posted: Fri Nov 17, 2017 12:18 pm
by ydna1977
Hi ,
Headspace 7697A carries out equilibration , pressurization and injection down transfer line
HP6890 then starts oven program (lights on HP6890 turn on)
Empower 3 sits on status 'Preparing - Sample for injection'
No trigger to collect and acquire data
Lace boxes all working and pinging (IP addresses ok) and you can monitor baseline and detector output ok
Any thoughts ? Thanks
Re: Headspace 7697A with 7890 - No trigger to acquire
Posted: Fri Nov 17, 2017 9:59 pm
by dlbenach
I've never had this happen to me, but I have 1 thought. In the Empower instrument method is "None" selected for the Injector? That's how we set up our methods. I was thinking if an injector was selected maybe that could cause this problem?????
Posted: Fri Nov 17, 2017 10:18 pm
by carl.nott
Is the GC actually running or is it waiting for something? Try seeing if hitting the 'Prep Run' button makes everything work. If it does you need to enable 'Auto Prep Run' in the GC Config.
Posted: Mon Nov 20, 2017 8:16 am
by ydna1977
Hi ,
Thanks for your reply
Yes - None is selected as injector type
Hitting prep run makes no difference
I will have a look at auto prep run
Any further thoughts let me know
Could it be the LACE box ?!
Posted: Thu Nov 30, 2017 1:41 pm
by ydna1977
Hi ,
Still no forward with this
When the 7697A headspace injects the status time on GC 6890 immediately starts counting upwards (Time = 0.xyz) so injection is recognised from HS to GC
Empower 3 remains on 'waiting on injection'
Any thoughts much appreciated
Posted: Thu Nov 30, 2017 5:15 pm
by dlbenach
What if you created a method where there is no injector, no headspace, and the Channel is set to TEST PLOT. When you load this method Empower will Wait for Injection. At this point, press the START button on the GC and see if Empower starts recording the TEST PLOT chromatogram. This will tell you if the GC is communicating with Empower.
Posted: Fri Dec 01, 2017 7:10 am
by ydna1977
Hi ,
I tried a manual injection ,, nor quite as you detailed above and Empower collected the data
I did notice configuration on back ALS injector
What should this be set at ? Does this come into play with remote control ?
Current options
No Headspace
Headspace 7697A (12 Vial)
Headspace 7697A (111Vial)
But when I select Headspace 7697A (111Vial) the GC out put tries to communicate with Empower and counts up
Downloading method info from Empower to GC and HS not a problem
Posted: Fri Dec 01, 2017 1:58 pm
by dlbenach
Sounds like all the instruments are communicating with each other. Empower can download the method to the GC and headspace. Empower can also monitor the baseline. The headspace triggers the GC to start. The GC triggers Empower to collect data when done manually. I think you are on the right track looking at the method and how the instruments are configured. The GC is waiting for something besides the headspace to trigger it to tell Empower to start collecting data. Maybe it's waiting for the back ALS injector?
Unfortunately, I do not have the exact same set up as you. I have Empower 3 controlling a 7890 with an Agilent G1888. I don't understand why you are seeing a back ALS (I'm assuming it's a 7693) injector. That could be the problem. Can you post a screen shot of the Settings screen of your Empower method.
Posted: Mon Dec 04, 2017 8:42 am
by ydna1977
Hi
Back ALS Injector is one of the options on the GC6890 (part of configuration)
I cant seem to get a screen shot onto the forum
Guide for adding picture doesn't work
Posted: Mon Dec 04, 2017 9:13 am
by dblux_
Hi
...Guide for adding picture doesn't work
Just post a link to a picture.
Posted: Mon Dec 04, 2017 10:10 am
by ydna1977
How do I do that ?
Posted: Mon Dec 04, 2017 10:25 am
by dblux_
How do I do that ?
Upload a picture to a server (eg. Imgur) and post url address here.
Posted: Mon Dec 04, 2017 10:54 am
by ydna1977
Posted: Mon Dec 04, 2017 11:27 am
by ydna1977
Some more photos,
Empower screen and GC/HS set up when running
https://imgur.com/a/hK3Pb
Posted: Mon Dec 04, 2017 9:35 pm
by dlbenach
From the photo it looks like the headspace is connected to the back inlet, and maybe a tower injector on the front inlet. Am I seeing that correctly? I'm thinking there is a problem with how the Empower method has the instruments configured, that's why I wanted to see a screen shot of the "Settings" tab screen of your Empower method. You posted the Run screen.
Have you ever ran a method with this configuration before?