Waters 2487 Detector

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

7 posts Page 1 of 1
I'm having a problem with the Waters 2487 dual wavelength detector. I keep getting an error message that reads “GRATING INITIATIZATION FAIL; NO HOME SENSORâ€
My suggested for cause
1. Maybe motor drive failure or
2. Possible home sensor fail.
Hello Kasiban,

Thank you for your comment. I will check into that & will follow up on this forum.

Cheers.

or junk in flow cell - pump something clean through it & power cycle it.
Thanks,
DR
Image

Thank you, DR. I will try your suggestion as well.

Hello again,

Here's what happened. We've installed a new lamp on the 2487, the sample and reference energy is about 20 nA at 245 nm. At the end of the manual calibration, we got a message that says "Peak was not found". What could be wrong? The flow cell was thoroughly cleaned, the lamp is self-aligned, there is nothing in the corvette. Could the energy level be too low or is there any other optical alignment option?

Hello iris.s7one

Have you re-calibrate the detector after install new lamp?

If done, go to next question

Have you change the detector's battery?
7 posts Page 1 of 1

Who is online

In total there are 32 users online :: 1 registered, 0 hidden and 31 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 31 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry