I run a Hewlitt-Packard 6890 GC with a 5793 MS. I see signigicant peaks in my solvent runs (we use ACS grade hexane as a solvent). I have cleaned the inlet, and the ion source, replaced the inlet liner, and base seal, and do not see these peaks when I run the method without making an injection. The air and water check for the system comes up clean (18/69 = 0.31, 28/69 = 1.12, 32/69 = 0.22, 44/69 = 0.19).

If I concentrate the hexane under a stream of nitrogen, and then inject it, the peak areas increase significantly, but I purchased a new lot of hexane, and the problem didn't go away.

I'm at a loss. Where could the contaminants be coming from?

I assume you are talking about chromatograohic peaks. What is their retention time? What is the GC temperature program? What is the elution temperature? What do the mass spectra for these chromatographic peaks look like?

:D
Regards;
David

O. David Sparkman
Consultant-At-Large
2 posts Page 1 of 1

Who is online

In total there are 9 users online :: 0 registered, 0 hidden and 9 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 9 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry