Sudden decrease in LC-MS peak area count
Posted: Mon Nov 06, 2017 1:38 am
Hi there,
We've recently validated a method, however, we now have an issue with the method whereby there is a drastic decrease in peak area count for the internal standard and analyte. This was not observed during the validation process.
During the validation, the ISTD peak count in matrix remained constant around 250,000 to 350,000. However, it has now decreased from 300,000 to 8,000 over the last few days. We initially thought it might have been a column degradation issue, however, this has been ruled out as it's observed in the new column as well.
We are thinking that it's most likely due to matrix effect as we had just tested 10 injections of matrix extracted sample followed by 20 injections of an aqueous sample (no matrix). For the matrix extracted sample, the ISTD peak count dropped from 55000 to 8000, and upon injection of the aqueous sample, a reverse trend is observed with the peak count increased back from 100000 to 300000.
Any ideas why this might be especially when we did not observe this issue throughout the validation runs (> 10 runs, with >1000 injections in total)? We're really stumped as to why this might be the case.
Thanks!
Thanks!