PDA detectors?
Posted: Sun Nov 05, 2017 6:10 am
by chemgc779
Hi, can someone please explain the purpose of purging PDA detectors - is it to flush out any residual contaminants/samples from previous run before starting a new run?
As far as I know, the detector should be purged for max. 30 minutes before autozeroing it before purging again.
Thank you.
Re: PDA detectors?
Posted: Sun Nov 05, 2017 11:36 am
by EmpowersBane
Do you mean as part of a start-up procedure? I use a PDA detector all the time and at the end of my run I flush 90:10 Water:Organic through the system to remove buffer, then 100% Organic for 2 hours before shutting off. Then at the start of the system I light up the lamp for at least 30 minutes before activating run and equilibrate the system with the starting conditions of my mobile phase.
When I get stray peaks or a noisy baseline and I can eliminate the check valves, contamination in my mobile, diluent or standards, then I perform an acid wash of 30% Phosphoric Acid on the entire system and that cleans off any residual drug or rubbish in the flow cell and surrounding connections.
Re: PDA detectors?
Posted: Sat Nov 11, 2017 10:54 pm
by Multidimensional
Perhaps you are confused and writing about the wrong type of detector. We do not purge PDA's. We do purge RID (Refractive Index Detector).
Re: PDA detectors?
Posted: Tue Nov 21, 2017 1:41 pm
by chuong08h2a.vn
Do you mean as part of a start-up procedure? I use a PDA detector all the time and at the end of my run I flush 90:10 Water:Organic through the system to remove buffer, then 100% Organic for 2 hours before shutting off. Then at the start of the system I light up the lamp for at least 30 minutes before activating run and equilibrate the system with the starting conditions of my mobile phase.
When I get stray peaks or a noisy baseline and I can eliminate the check valves, contamination in my mobile, diluent or standards, then I perform an acid wash of 30% Phosphoric Acid on the entire system and that cleans off any residual drug or rubbish in the flow cell and surrounding connections.
30%.. I think it should be 17,5%.can u explain to me? tks
Re: PDA detectors?
Posted: Wed Nov 22, 2017 7:05 pm
by Multidimensional
Please do not wash down the system with acid. What does this have to do with "purging" ?? I think you need to speak with someone at your company who is experienced in chromatography.