pre-mixed or non-pre-mixed

Posted: Fri Dec 15, 2006 5:52 pm
by ym3142
This is a general question. Recently we found in several cases bubble was generated through the solvent inlet filter when the solvent was a premixed mobile phase (typically 80 Buffer and 20 ACN). A brand new filter or re-cleaned filter will solve the problem. But I never see bubble was generated in 100% organic or 100% buffer. So is the mixed solvent more likely to generate the bubble or I just conincidently had some dirty filters?

Posted: Fri Dec 15, 2006 6:29 pm
by Mark Tracy
Air is much more soluble in pure organic solvents than in water. When you mix them you get a supersaturated solution of air. It spontaneously outgasses for a long time. You could briefly degas the mixture with vacuum/sonication if you like.

Posted: Sat Dec 16, 2006 5:23 am
by ym3142
Thanks, Mark

Posted: Sun Dec 17, 2006 12:36 am
by danko
A single remark: It’s true that one doesn’t see air bubbles in 100 % organic and to some extend in 100 % aqueous mobile phases. The reason is that the air bubbles are released upon mixing these i.e. in the solvent manager. That’s the last place I wish to deal with air :wink:

Best Regards

Posted: Sun Dec 17, 2006 4:50 am
by ym3142
thanks, Danko