appears to handle 100% acetonitrile better than some other brand pumps - which may encounter sticking valves, unless water or water blends are regularly pumped in the acetonitrile channel.
I’m just curious about the reason for this malfunctioning of the valves, when 100 % ACN is pumped through some pumps. Does something dissolve?
Not as far as I'm aware, but I don't know the details. It's been suggested that the stickiness is mainly due to active seat surfaces binding some molecules - producing a very smooth surface that no longer has microscopic grinding marks. The smooth surface permits a thin film to hold the valve shut just by surface tension.
For some reason, acetonitrile or acetontrile/water cause the problem.
In my limited experience, occasionally flushing the pump with water seems to prevent the problem, but why is a mystery to me, although some obvious speculations appear....
The problem was initially defined as occuring with ruby ball valves and sapphire seats, as in many early HPLC pumps. There is a Spectra-Physics US patent ( 4,862,907 to Ledtje , et al. September 5, 1989 ) for a check valve claiming to solve the problem. They investigated the water/acetonitrile problem using SEM and FTIR, and finally offered up ceramic valves as the solution.
You can read the patent for free at the USPTO www site. Their abstract claimed:-
" An improved liquid flow check valve using a ball and seat wherein the seat is made of sintered ceramic. The improved valve will operate in an environment of acetonitrile and water without sticking shut, thereby providing more stable liquid flow rates in liquid chromatography systems. "
Hope this is interesting.
Bruce Hamilton