LC-MS/MS pH=10.2 Xbridge, m/z=141 contamination
Posted: Thu Dec 14, 2006 3:48 pm
Hi,
I have some problems with contamination on my Waters Micro LC-MS-MS system, I hope somebody out there have some ideas to help me!
HPLC method
Solvent A: 5mM ammoniumformate in H2O adjustet to pH=10.2 with ammoniahydroxide, 5% acetonitril.
Solvent B: Acetonitrile.
A gradient program is run. Column: Waters Xbridge.
I am analyzing amitriptyline and metabolites, there is a nice separation because of the basic pH.
Initially I used a Phenomenex Gemini column, but they allways died after ~500 injections with peak tailing and rising backpressure. Switching to Waters Xbridge column have so far given me 1000 injections and there is still nice peaks.
The problem: In MS-SCAN mode I have a huge contamination at m/z=141 - and also a peak at m/z=59. Running pure metanol into the MS from a syringe the m/z=141 peak is still there. Cleaning ths MS (Cone, extrator cone and hexapole/quadrupole) helped the situation. But when I used my HPLC method again the contamination got back. Where does this conatamination originate from? (and what is it?)
The most obvious reason is breakdown products from the column, origination from the use of the basic solvent A. I have tried different basic solvents but the described solvent is way the best for my analytes. Ammonium bicarbonate was not good, giving irreproducible results. Other suggestions for a basic solvent for LC-MS - would also be nice!
If you have experience running LC-MS at basic pH you are more than welcome to share your knowledge (column lifetime, solvent selection and other problems)!
Best regards Jakob Jornil
I have some problems with contamination on my Waters Micro LC-MS-MS system, I hope somebody out there have some ideas to help me!
HPLC method
Solvent A: 5mM ammoniumformate in H2O adjustet to pH=10.2 with ammoniahydroxide, 5% acetonitril.
Solvent B: Acetonitrile.
A gradient program is run. Column: Waters Xbridge.
I am analyzing amitriptyline and metabolites, there is a nice separation because of the basic pH.
Initially I used a Phenomenex Gemini column, but they allways died after ~500 injections with peak tailing and rising backpressure. Switching to Waters Xbridge column have so far given me 1000 injections and there is still nice peaks.
The problem: In MS-SCAN mode I have a huge contamination at m/z=141 - and also a peak at m/z=59. Running pure metanol into the MS from a syringe the m/z=141 peak is still there. Cleaning ths MS (Cone, extrator cone and hexapole/quadrupole) helped the situation. But when I used my HPLC method again the contamination got back. Where does this conatamination originate from? (and what is it?)
The most obvious reason is breakdown products from the column, origination from the use of the basic solvent A. I have tried different basic solvents but the described solvent is way the best for my analytes. Ammonium bicarbonate was not good, giving irreproducible results. Other suggestions for a basic solvent for LC-MS - would also be nice!
If you have experience running LC-MS at basic pH you are more than welcome to share your knowledge (column lifetime, solvent selection and other problems)!
Best regards Jakob Jornil