Verification of compendial methods
Posted: Fri Oct 27, 2017 10:50 am
by danor285
Hello.
What tests of an analytical validation do you perform?
Re: Verification of compendial methods
Posted: Tue Oct 31, 2017 11:33 am
by Khanom
Hi Danor,
I am not sure of your type of analysis, but in general method validation parameters of HPLC/IC/UHPLC methods cover the following areas:-
System Suitability Tests:-
• Capacity factor
• Precision/ Injection repeatability
• Relative retention
• Resolution
• Tailing factor
• Theoretical plate number
and:-
• Accuracy
• Detection Limit and Quantitation Limit
• Linearity
• Precision
• Repeatability
o Injection Repeatability
o Analysis Repeatability
o Intermediate Precision
o Reproducibility
• Range
• Recovery
• Robustness
• Sample Solution Stability
• Specificity / selectivity
Posted: Tue Oct 31, 2017 2:38 pm
by Consumer Products Guy
I think that the original poster is asking about "verification" of compendial methods to use in his facility, which is different from validation requirements when developing a new method.
You and your QA should investigate and write up as an SOP, and then follow that. For example, when we set up trace ethylene glycol and diethylene glycol USP test for glycerin, we just followed the USP procedure as written, so only nominal testing was required.