i have no peak

Posted: Thu Oct 26, 2017 6:16 am
by chemgirl00
hi :) i have problems about my peaks. i injected system suitability solution first, everything is okay and started sample set. but couldn't dedect any peak. system is gradient and flow: 0,27. i checked everything and as same as methods.
waiting your suggestions..

Re: i have no peak

Posted: Thu Oct 26, 2017 7:28 am
by bunnahabhain
I suggest that you give more information:
What system, what system suitability samples, what eluent, what detector, ...

Re: i have no peak

Posted: Thu Oct 26, 2017 11:32 am
by chemgirl00
System HP 1100, column;C18,150 x 2.0 mm 3µm, Flow:0,27 and gradient, Detection: 200nm,i connected two columns serial.system conditions like this. when i saw peaks started immediately set but where is gone my peak, i didn' t know :)

Re: i have no peak

Posted: Thu Oct 26, 2017 11:33 am
by BostonFSE
Possible reasons for no peak:
your purge valve on the pump is open
You selected the wrong wavelength
Your sample volume in vial is too low
You have air in your system (purge all lines for 5 min each)
Your method run time is too short, peak is still on column

Re: i have no peak

Posted: Thu Oct 26, 2017 11:52 am
by chemgirl00
purge valve off i'm sure. only not sample, and standard i didn't see peak. purged my system. system was high pressure. i tried analysis shimadzu system but didn't work because of high pressure. this peaks wasn't seen; because of high pressure?